The Nanoscale World

Attenuation of Input 1 on NSV Controller?

rated by 0 users
Answered (Verified) This post has 1 verified answer | 1 Reply | 2 Followers

Top 100 Contributor
10 Posts
Points 112
rwfriddle posted on Tue, Sep 25 2012 2:13 PM

Hello-

Is is possible to attenuate the min/max range of the ADCs linked to the front panel BNCs of the Nanoscope V controller? Thanks.

-Ray

  • | Post Points: 12

Answered (Verified) Verified Answer

Top 100 Contributor
8 Posts
Points 94
Bruker Employee
Verified by rwfriddle

Ray,

The ADCs that are linked to the front panel have a fixed input range of ~+/-12V. These ADCs give true 16 bit resolution -- so, every LSB accounts for ~0.4 mV. If you need better resolution, I would recommend gaining up your signal before the front panel BNCs.

Hope this helps,

Fritz

  • | Post Points: 11
Page 1 of 1 (2 items) | RSS
Copyright (c) 2011 Bruker Instruments