Hi,
I'm still trying to find how to read header datas in AFM force files and in particular how to find the Z scale (to find Voltage values). Then for the conversion from Voltage to nm, it is easy.
In that file the expected conversion factor is 0.06615 V/LSB for z movement. Where can I find such values or a mean to obtain such value in my file.
enclosed the header of my file of data
thanks in advance, sincerely
Hervé
\*Force file list
\Version: 0x07100000
\Date: 03:10:26 PM Fri Feb 10 2012
\Start context: FOL
\Data length: 40960
\Text:
\History:
\Navigator note:
\Engage X Pos:
\Engage Y Pos:
\*Equipment list
\Description: Dimension V
\Microscope: Dimension V
\Extender: None
\Sensor: None
\Vision: Generic Frame Grabber
\Scanner file: hybrid-279.scn
\*Scanner list
\Scanner type: Dimension
\Serial Number: Hybrid-279
\Piezo size: G
\File name: hybrid-279.scn
\Retracted offset der: 4
\Extended offset der: 16.8
\Allow rotation: Allow
\Piezo cal: 440
\X sensitivity: 144.228
\X derate: 0.2103
\X mag: 1.2
\X mag1: 0.75
\X arg: 3.25
\Fast arg derate: 0.001
\X round: 0
\Orthogonality: 0.00349066
\Y sensitivity: 149.224
\Y derate: 0.1904
\Y mag: 1.22
\Y mag1: 0.75
\Y arg: 2.85
\Slow arg derate: 0.001
\X slow sensitivity: 157.412
\X slow derate: 0.2627
\Y fast sensitivity: 142.4
\Y fast derate: 0.2
\X slow-fast coupling: 0.4962
\X slow-fast coupling derating: 0
\Y slow-fast coupling: 0.3992
\Y slow-fast coupling derating: -0.0001124
\Fast cal freq: 2.44141
\Slow cal freq: 9.54
\Xs-Yf coupling: 0.02632
\Xs-Yf coupling derating: -0.0001136
\Ys-Xf coupling: 0.1387
\Ys-Xf coupling derating: 0
\X offset sens: 300
\Y offset sens: 300
\Max scan size: 110
\Z min: -220
\Z max: 110
\Z engage position: 0
\Z to X coupling: 15
\Z to Y coupling: 0
\OptoXY Scanner: Yes
\OptoXY x sensitivity: 318.9
\OptoXY y sensitivity: 291.6
\OptoXY orthogonality: -0.0261799
\OptoXY x sens fitted: 318.9
\OptoXY y sens fitted: 291.6
\OptoXY orthog fitted: -0.0261799
\OptoXY scan angle cor: -0.00959931
\OptoXY pgain/igain: 0.5
\Optoxy FitVersion: V1
\OptoXY xsenscal_x: 0
\OptoXY xsenscal_y: 0
\OptoXY xsenscal_xy: 0
\OptoXY xsenscal_xx: 0
\OptoXY xsenscal_yy: 0
\OptoXY xsenscal_xxy: 0
\OptoXY xsenscal_xyy: 0
\OptoXY xsenscal_xxx: 0
\OptoXY xsenscal_yyy: 0
\OptoXY xsenscal_xxyy: 0
\OptoXY xsenscal_xxxy: 0
\OptoXY xsenscal_xyyy: 0
\OptoXY xsenscal_xxxx: 0
\OptoXY xsenscal_yyyy: 0
\OptoXY ysenscal_x: 0
\OptoXY ysenscal_y: 0
\OptoXY ysenscal_xy: 0
\OptoXY ysenscal_xx: 0
\OptoXY ysenscal_yy: 0
\OptoXY ysenscal_xxy: 0
\OptoXY ysenscal_xyy: 0
\OptoXY ysenscal_xxx: 0
\OptoXY ysenscal_yyy: 0
\OptoXY ysenscal_xxyy: 0
\OptoXY ysenscal_xxxy: 0
\OptoXY ysenscal_xyyy: 0
\OptoXY ysenscal_xxxx: 0
\OptoXY ysenscal_yyyy: 0
\OptoXY orthocal_x: 0
\OptoXY orthocal_y: 0
\OptoXY orthocal_xy: 0
\OptoXY orthocal_xx: 0
\OptoXY orthocal_yy: 0
\ClosedLoop xy threshold: 300
\ClosedLoop x igain: 15
\ClosedLoop x pgain: 7.5
\ClosedLoop y igain: 15
\ClosedLoop y pgain: 7.5
\Z range sens: 25.3603
\OptoZ Scanner: Yes
\Z sensor sens: 782.98
\Z sensor gain factor: 1.5
\Z CL range factor: 0.9
\Z CL IGain: 49
\Z CL PGain: 77
\Full Range Sensor Gain: 0.7
\@Sens. Zsens: V 23.61771 nm/V
\@Sens. SECPMSens: V 100.0000 mV/V
\@Sens. CurrentSens: V 10.00000 nA/V
\*Ciao scan list
\Parameter select: Main
\Operating mode: Force
\Tip Serial Number:
\Scan Size: 1102.45 nm
\X Offset: 2131.67 nm
\Y Offset: 7151.28 nm
\Rotate Ang.: 0
\Samps/line: 256
\Lines: 256
\Y disable: Enabled
\Aspect Ratio: 1:1
\Bidirectional Scan: Disabled
\Scan line shift: 0
\Scan Rate: 1.0016
\CD Mode: Disabled
\PrpGain: 0.5
\IntGain: 0.5
\PIDErr: 0.0152588
\tappFilterCutoff: 100
\dithFilterCutoff: 100
\LocHighLowDistance: 200
\SlopeOutOffset: 0.5
\SlopeOutLimit: 0.5
\PointLag: 100
\RcrsvAvg: 100
\AvTipState: 100
\microstepSize: 5
\zxRatio: 10
\dataStepAdj: 1
\debugData: Standard
\Amplitude X: 0
\Amplitude Y: 0
\Phase X: 0
\Phase Y: 0
\spare0: 0
\spare1: 0
\spare2: 0
\spare3: 0
\spare4: 0
\spare5: 0
\Tip Velocity: 2.20843
\Minimum scan rate: 0.1
\Lift rate: 4
\X drift: 0
\Y drift: 0
\Step XY size: 300
\Cycles: 10
\Step XY period: 0.005
\Step size: 3
\Units: Metric
\Color table: 2
\Scope dualtrace: Dual
\Auto X Sep: 0
\Auto Y Sep: 0
\Auto pattern: Linear
\Auto number: 2
\Capture direction: Up
\Capture prelines: 50
\Engage Setpoint: 1
\Min int. gain: 0.2
\Parameter Update Retract: Enabled
\Drive feedback: Disabled
\Drive time: 0
\Drive setpoint: 1
\Drive gain: 80
\Profiler profiling: Disabled
\Profiler length goal: 500
\Profiler length actual: 0.001
\Profiler speed: 1
\Profiler scan time: 0.001
\Profiler resolution: 0.001
\Profiler x position: 0
\Profiler y position: 0
\Profiler z position: 0
\Profiler profiler position: 0
\Profiler tilt angle: 0
\Profiler direction: Forward
\Profiler axis: Aux
\Profiler ramp-up ideal-distance multiplier: 25
\Profiler ramp-up total-distance minimum: 2
\Profiler ramp-up total-distance goal: 10
\Profiler backlash-removal distance: 10
\Profiler backlash-removal speed: 33
\Profiler num profile axis display points 3D: 512
\Gain start: 32
\Gain end: 192
\Gain incr: 32
\Pro. Gain factor: 1.5
\Max shift: 0.05
\Lines/gain: 3
\Gain offset: 0.9
\Auto Gain: Disabled
\Start size: 13199.8 nm
\Pass 2 size: 10000 nm
\Pass 3 size: 5000 nm
\Pass 4 size: 2000 nm
\Feature scan angle: 0
\Feature Scan rate: 1
\Feature step size: 0.138889
\Feature setpoint: 0.8
\Feature threshold: 2000
\Pass delay: 5
\Z offset: 0
\Z tolerance: 3237.22
\Engage X offset: 0 nm
\Engage Y offset: 0 nm
\Engage step size: 0.138889
\Engage Z sens.: 12.5
\Engage Z avg.: 10
\Engage Z delay: 0.5
\Height engage: Disabled
\Current Sensitivity: 1 pA/V
\Extended Current sens: 1 pA/V
\Bias limit: 12 V
\AFM Current Sensitivity: 1 nA/V
\SECPM sens: 100 mV/V
\Nanopot gain: 1x
\Z Range: 8.3689
\HF current: 1
\HF freq.: 100
\HF power: -30
\HF RF: Off
\HF AM: Off
\HF lower limit: -20
\HF upper limit: 0
\HF int. gain: 5
\HF prop. gain: 5
\HF engage: Disabled
\HF current capture: 0
\HF display: Current
\MR preamp gain: 1000
\MSM bias voltage: 0
\MSM resistance: 150
\MSM bias delay: 1
\MSM test current: 3
\MSM max sensor resist: 200
\MSM min sensor resist: 10
\MSM enable: Enabled
\MSM Probe Delay: 2
\Engage MRR: -1
\Scan MRR: -1
\Strip chart : Disabled
\Strip Chart Rate: 500
\Strip Chart Size: 100
\XY Closed Loop: Off
\OptoXY Do Linear: Off
\Z Closed Loop: On
\Analog2 High Voltage: 0
\Analog22 High Voltage: 0
\Output 1 Data Type: Off
\Output 2 Data Type: Off
\@InterleaveList: S [InterleaveOffMode] "Disabled"
\@LinearizeList: S [LinearizeOffMode] "modeLinearizeOff"
\@MicroscopeList: S [TMMode] "Tapping"
\@LP Deflection: S [] "Enabled"
\@LP TM Deflection: S [] "Enabled"
\@LP Current: S [] "Enabled"
\@LP Vertical Deflection: S [] "Enabled"
\@Low Pass Frequency: S [] "Enabled"
\@LP Friction: S [] "Enabled"
\@PicoAnglerPollOnOff: S [PicoAnglerPollOff] "Disabled"
\@ThermalTuneMaxFreq: S [ThermalTuneData500K] "1 - 100 KHz"
\@TipBiasCntrlList: S [TipBiasCntrlGround] "Ground"
\@Sample Bias Control: S [] "Ground"
\@HSADCMuxOnOff: S [HSADCMuxOff] "modeHSADCMuxOff"
\@HsdcChanADataType: S [HsdcDataTypeDeflection] "Vertical Deflection"
\@HsdcChanBDataType: S [HsdcDataTypeLateral] "Lateral Deflection"
\@FPGA Z Feedback: S [] "On"
\@FastScan List: S [FastScanOff] "Disabled"
\@FastZFeedback List: S [DigitalZFastScan] "Digital"
\@TR Mode: S [] "Disabled"
\@SCM Feedback: S [] ""
\@TUNAList: S [TUNAOffMode] "Open"
\@Feedback: S [TUNAOffMode] "Open"
\@4:SPMFeedbackList: S [SPMFb] "TM Deflection"
\@3:SPMFeedbackList: S [SPMFb] "Amplitude"
\@4:Drive1Routing: S [] "Tapping Piezo"
\@3:Drive1Routing: S [] "Tapping Piezo"
\@4:Drive1Monitor: S [] "BNC 1 (DDS1)"
\@3:Drive1Monitor: S [] "BNC 1 (DDS1)"
\@2:MainLockInEnableContactMode: S [LockIn1Disabled] "Disabled"
\@3:MainLockInEnableContactMode: S [LockIn1Disabled] "Disabled"
\@4:MainLockInEnableNonContactMode: S [LockIn1Enabled] "Enabled"
\@3:MainLockInEnableNonContactMode: S [LockIn1Enabled] "Enabled"
\@4:MainLockInSource: S [] "Vertical"
\@3:MainLockInSource: S [] "Vertical"
\@4:Zmod: S [] "Disabled"
\@3:Zmod: S [] "Disabled"
\@4:InputFbList: S [] "Off"
\@3:InputFbList: S [] "Off"
\@2:TR Mode: S [] "Disabled"
\@3:TR Mode: S [] "Disabled"
\@2:TRCouplingCheck: S [] ""
\@3:TRCouplingCheck: S [] ""
\@4:Drive2Routing: S [] "Null"
\@3:Drive2Routing: S [] "Null"
\@4:Drive2Monitor: S [] "Not Available"
\@3:Drive2Monitor: S [] "Not Available"
\@4:LockIn2DDS2Enable: S [LockIn2Disabled] "Disabled"
\@3:LockIn2DDS2Enable: S [LockIn2Disabled] "Disabled"
\@2:DDS2HarmonicList: S [] "1"
\@3:DDS2HarmonicList: S [] "1"
\@4:LockIn2Source: S [] "Horizontal"
\@3:LockIn2Source: S [] "Vertical"
\@2:PiezoACBiasCtrlList: S [] ""
\@3:PiezoACBiasCtrlList: S [] ""
\@2:SCM Feedback: S [] ""
\@3:SCM Feedback: S [] ""
\@4:Drive3Routing: S [] "Null"
\@3:Drive3Routing: S [] "Null"
\@4:Drive3Monitor: S [] "Not Available"
\@3:Drive3Monitor: S [] "Not Available"
\@4:LockIn3DDS3Enable: S [LockIn3Disabled] "Disabled"
\@3:LockIn3DDS3Enable: S [LockIn3Disabled] "Disabled"
\@4:LockIn3Source: S [] "Front Panel"
\@3:LockIn3Source: S [] "Vertical"
\@4:HsdcChanCDataType: S [HsdcDataTypeOff] "Off"
\@3:HsdcChanCDataType: S [HsdcDataTypeOff] "Off"
\@4:HsdcChanDDataType: S [] "Height"
\@3:HsdcChanDDataType: S [HsdcDataTypeOff] "Off"
\@4:HsdcTriggerChan: S [] "Height"
\@3:HsdcTriggerChan: S [] "Height"
\@Sens. ECBiasSens: V 1000.000 mV/V
\@Sens. DeflSens: V 43.21322 nm/V
\@Sens. AFMSetFastScanSens: V 1.000000 V/V
\@Sens. ZsensSens: V 78.29761 nm/V
\@Sens. ZPicoSens: V 1000.000 nm/V
\@Sens. tunaSens: V 10.00000 pA/V
\@Sens. biassens: V 1.000000 V/V
\@Sens. testsens: V 1.000000 V/V
\@Sens. Raw Deflection: V 1.000000
\@Sens. Raw Deflection Gain: V 1.000000
\@Sens. Counter 1: V 1.000000
\@Sens. Counter 2: V 1.000000
\@Sens. Freq. 1: V 1.000000
\@Sens. Freq. 2: V 1.000000
\@Sens. Amplitude: V 1.000000
\@Sens. Amplitude Error: V 1.000000
\@Sens. Phase: V 1.000000
\@Sens. Phase Error: V 1.000000
\@Sens. Friction: V 1.000000
\@Sens. TM Deflection: V 1.000000
\@Sens. TM Deflection Error: V 1.000000
\@Sens. Inphase: V 1.000000
\@Sens. Quadrature: V 1.000000
\@Sens. Fast Z Analog: V 1.000000
\@Sens. Fast Z: V 1.000000
\@Sens. Fast RMS: V 1.000000
\@Sens. Fast Deflection Analog: V 1.000000
\@Sens. Fast Deflection Error: V 1.000000
\@Sens. Xsensor: V 1.000000
\@Sens. Ysensor: V 1.000000
\@Sens. Frequency: V 1.000000
\@Sens. Potential: V 1.000000
\@Sens. TR Amplitude: V 1.000000
\@Sens. TR Amplitude Error: V 1.000000
\@Sens. TR Phase: V 1.000000
\@Sens. Amplitude2: V 1.000000
\@Sens. Amplitude 2: V 1.000000
\@Sens. dC/dV Amplitude: V 1.000000
\@Sens. dC/dV Amplitude Error: V 1.000000
\@Sens. dC/dV I: V 1.000000
\@Sens. dC/dV Q: V 1.000000
\@Sens. dC/dV Phase: V 1.000000
\@Sens. Feedback Bias: V 1.000000
\@Sens. Resistance (Ohms): V 1.000000
\@Sens. Resistance (V): V 1.000000
\@Sens. Input1: V 1.000000
\@Sens. Input2: V 1.000000
\@Sens. Input3: V 1.000000
\@Sample period: V (0.1000000 us/LSB) 10.00000 us
\@Z center: V [Sens. Zsens] (0.006713867 V/LSB) 55.00000 V
\@1:AmplitudeLimit1: V (0.0000009313 mV/LSB) 2000.000 mV
\@1:Z Limit: V [Sens. Zsens] (0.006713867 V/LSB) 330.0000 V
\@1:TRAmplitudeLimit: V (0.0000009313 mV/LSB) 2500.000 mV
\@1:SCMAmplitudeLimit: V (0.0000009313 mV/LSB) 4000.000 mV
\@2:AmplitudeLimit1LockIn1: V (0.0000009313 mV/LSB) 4000.000 mV
\@2:AmplitudeLimit1LockIn2: V (0.0000009313 mV/LSB) 4000.000 mV
\@2:DeflectionLimitLockIn3LSADC1: V (0.0000000057 V/LSB) 24.57600 V
\@2:DeflectionLimit: V (0.0000000057 V/LSB) 24.57600 V
\@2:TMDeflectionLimitLockIn3LSADC1: V (0.0000000057 V/LSB) 24.57600 V
\@2:TMDeflectionLimit: V (0.0000000057 V/LSB) 24.57600 V
\@2:HSADC2GainLockIn2: V (0.0000009313 mV/LSB) 4000.000 mV
\@2:HSADC2GainLockIn1: V (0.0000009313 mV/LSB) 4000.000 mV
\@2:LSADC2Gain: V (0.0000000057 V/LSB) 24.57600 V
\@2:LSADC9Gain: V (0.0000000057 V/LSB) 24.57600 V
\@Lift Start Height: V [Sens. Zsens] (0.002861097 V/LSB) 0 V
\@Lift Scan Height: V [Sens. Zsens] (0.002861097 V/LSB) 4.506534 V
\@Drive height: V [Sens. Zsens] (0.002861097 V/LSB) 0 V
\@2:AFMSetDeflection: V (0.0003662109 V/LSB) -0.7000000 V
\@3:AFMSetDeflection: V (0.0003662109 V/LSB) 0 V
\@2:TMSetAmplitude: V (0.03051758 mV/LSB) 263.3972 mV
\@3:TMSetAmplitude: V (0.03051758 mV/LSB) 0 mV
\@2:TMSetDeflection: V (0.0003662109 V/LSB) 4.455496 V
\@3:TMSetDeflection: V (0.0003662109 V/LSB) 0 V
\@2:TMSetFastScan: V (0.0003662109 V/LSB) 0 V
\@3:TMSetFastScan: V (0.0003662109 V/LSB) 0 V
\@2:AFMSetFastScanAnalog: V [Sens. AFMSetFastScanSens] (0.0003662109 V/LSB) 0 V
\@3:AFMSetFastScanAnalog: V [Sens. AFMSetFastScanSens] (0.0003662109 V/LSB) 0 V
\@2:AFMSetFastScanDigital: V [Sens. AFMSetFastScanSens] (0.0003662109 V/LSB) 0 V
\@3:AFMSetFastScanDigital: V [Sens. AFMSetFastScanSens] (0.0003662109 V/LSB) 0 V
\@2:STMSetCurrent: V [Sens. CurrentSens] (0.0003750000 V/LSB) 0.1000000 V
\@3:STMSetCurrent: V [Sens. CurrentSens] (0.0003750000 V/LSB) 10.00000 V
\@2:TRSetAmplitude: V (0.01525879 mV/LSB) 0 mV
\@3:TRSetAmplitude: V (0.01525879 mV/LSB) 0 mV
\@2:SetTRVertDefl: V (0.0003662109 V/LSB) 0 V
\@3:SetTRVertDefl: V (0.0003662109 V/LSB) 0 V
\@2:SECPMSetPotential: V [Sens. SECPMSens] (0.0003750000 V/LSB) 2.000000 V
\@3:SECPMSetPotential: V [Sens. SECPMSens] (0.0003750000 V/LSB) 2.000000 V
\@2:TMSetPhase: V (0.005493164 ∫/LSB) 0 ∫
\@3:TMSetPhase: V (0.005493164 ∫/LSB) 0 ∫
\@2:CantFrequency: V (0.0000116415 kHz/LSB) 60.16912 kHz
\@3:CantFrequency: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:CantFrequencyContactMode: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@3:CantFrequencyContactMode: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:CantDrive: V (0.3051758 mV/LSB) 0 mV
\@3:CantDrive: V (0.3051758 mV/LSB) 0 mV
\@2:CantDriveContactMode: V (0.3051758 mV/LSB) 0 mV
\@3:CantDriveContactMode: V (0.3051758 mV/LSB) 0 mV
\@2:CantDriveNulled: V (0.3051758 mV/LSB) 0 mV
\@3:CantDriveNulled: V (0.3051758 mV/LSB) 0 mV
\@2:DriveDCOffsetTappingMode: V (0.0003051758 V/LSB) 0 V
\@3:DriveDCOffsetTappingMode: V (0.0003051758 V/LSB) 0 V
\@2:DriveDCOffsetNonTapping: V (0.0003051758 V/LSB) 0 V
\@3:DriveDCOffsetNonTapping: V (0.0003051758 V/LSB) 0 V
\@2:ReferenceFrequencyTappingMode: V (0.0000116415 kHz/LSB) 60.16912 kHz
\@3:ReferenceFrequencyTappingMode: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:ReferenceFrequency1: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@3:ReferenceFrequency1: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:CantPhase: V (0.005493164 ∫/LSB) 126.5021 ∫
\@3:CantPhase: V (0.005493164 ∫/LSB) 0 ∫
\@2:CantPhaseContactMode: V (0.005493164 ∫/LSB) 0 ∫
\@3:CantPhaseContactMode: V (0.005493164 ∫/LSB) 0 ∫
\@2:LockInBW1: V (0.001000000 kHz/LSB) 24.58895 kHz
\@3:LockInBW1: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@2:LockInBW1ContactMode: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@3:LockInBW1ContactMode: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@2:TRCantileverFrequency: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@3:TRCantileverFrequency: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:ReferenceFrequencyTRMode: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@3:ReferenceFrequencyTRMode: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:TROnAnalog3: V (0.0003051758 V/LSB) 10.00000 V
\@3:TROnAnalog3: V (0.0003051758 V/LSB) 10.00000 V
\@2:DriveFrequency2: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@3:DriveFrequency2: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:DriveAmplitude2: V (0.3051758 mV/LSB) 0 mV
\@3:DriveAmplitude2: V (0.3051758 mV/LSB) 0 mV
\@2:Drive2 DC Offset: V (0.0003051758 V/LSB) 0 V
\@3:Drive2 DC Offset: V (0.0003051758 V/LSB) 0 V
\@2:ReferenceFrequency2: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@3:ReferenceFrequency2: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:ReferenceFrequency2Harmonic: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@3:ReferenceFrequency2Harmonic: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:DrivePhase2: V (0.005493164 ∫/LSB) 0 ∫
\@3:DrivePhase2: V (0.005493164 ∫/LSB) 0 ∫
\@2:LockIn2BW: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@3:LockIn2BW: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@2:PRLockInBW: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@3:PRLockInBW: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@2:ACBiasAmpl: V (0.3051758 mV/LSB) 0 mV
\@3:ACBiasAmpl: V (0.3051758 mV/LSB) 0 mV
\@2:BiasFrequency: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@3:BiasFrequency: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:SCMLockInBW: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@3:SCMLockInBW: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@2:SCM Lock-In Phase: V (0.005493164 ∫/LSB) 0 ∫
\@3:SCM Lock-In Phase: V (0.005493164 ∫/LSB) 0 ∫
\@2:DC Bias: V (0.0003051758 V/LSB) 0 V
\@3:DC Bias: V (0.0003051758 V/LSB) 0 V
\@2:Capacitance Sensor Frequency: V (0.03204346 MHz/LSB) 880.0000 MHz
\@3:Capacitance Sensor Frequency: V (0.03204346 MHz/LSB) 880.0000 MHz
\@2:SCMIgain: V (0.001525879 1/LSB) 10.00000
\@3:SCMIgain: V (0.001525879 1/LSB) 10.00000
\@2:SCMPgain: V (0.001525879 1/LSB) 10.00000
\@3:SCMPgain: V (0.001525879 1/LSB) 10.00000
\@2:SCMFeedbackSetpoint: V (0.01525879 mV/LSB) 200.0000 mV
\@3:SCMFeedbackSetpoint: V (0.01525879 mV/LSB) 200.0000 mV
\@2:SCMFeedbackLimit: V (0.0003051758 V/LSB) 6.000000 V
\@3:SCMFeedbackLimit: V (0.0003051758 V/LSB) 6.000000 V
\@2:DC Sample Bias: V [Sens. biassens] (0.0003051758 V/LSB) 0 V
\@3:DC Sample Bias: V [Sens. biassens] (0.0003051758 V/LSB) 0 V
\@2:DriveFrequency3: V (0.0000001164 kHz/LSB) 0.005000000 kHz
\@3:DriveFrequency3: V (0.0000001164 kHz/LSB) 0.005000000 kHz
\@2:DriveAmplitude3: V (0.3051758 mV/LSB) 0 mV
\@3:DriveAmplitude3: V (0.3051758 mV/LSB) 0 mV
\@2:Drive3 DC Offset: V (0.0003051758 V/LSB) 0 V
\@3:Drive3 DC Offset: V (0.0003051758 V/LSB) 0 V
\@2:ReferenceFrequency3: V (0.0000001164 kHz/LSB) 0.005000000 kHz
\@3:ReferenceFrequency3: V (0.0000001164 kHz/LSB) 0.005000000 kHz
\@2:LockIn3BW: V (0.001000000 kHz/LSB) 0.005000000 kHz
\@3:LockIn3BW: V (0.001000000 kHz/LSB) 0.005000000 kHz
\@2:DC Test Bias: V [Sens. testsens] (0.0002929687 V/LSB) 0 V
\@3:DC Test Bias: V [Sens. testsens] (0.0002929687 V/LSB) 0 V
\@2:Sample Bias: V (0.3051758 mV/LSB) 0 mV
\@3:Sample Bias: V (0.3051758 mV/LSB) 0 mV
\@2:Tip Bias: V (0.0003662109 V/LSB) 0 V
\@3:Tip Bias: V (0.0003662109 V/LSB) 0 V
\@2:Analog3: V (0.0003051758 V/LSB) 0 V
\@3:Analog3: V (0.0003051758 V/LSB) 0 V
\@2:Analog4: V (0.0003051758 V/LSB) 0 V
\@3:Analog4: V (0.0003051758 V/LSB) 0 V
\@2:TR Drive Amplitude: V (0.3051758 mV/LSB) 0 mV
\@3:TR Drive Amplitude: V (0.3051758 mV/LSB) 0 mV
\@2:TR Lock-In Phase: V (0.005493164 ∫/LSB) 0 ∫
\@3:TR Lock-In Phase: V (0.005493164 ∫/LSB) 0 ∫
\@2:PRDriveAmplitude: V (0.3051758 mV/LSB) 0 mV
\@3:PRDriveAmplitude: V (0.3051758 mV/LSB) 0 mV
\@2:PRDriveFrequency: V (0.0000116415 kHz/LSB) 30.00000 kHz
\@3:PRDriveFrequency: V (0.0000116415 kHz/LSB) 0.5000000 kHz
\@2:PR Drive Phase: V (0.005493164 ∫/LSB) 0 ∫
\@3:PR Drive Phase: V (0.005493164 ∫/LSB) 0 ∫
\@2:DrivePhase3: V (0.005493164 ∫/LSB) 0 ∫
\@3:DrivePhase3: V (0.005493164 ∫/LSB) 0 ∫
\@2:SSRMSampleBias: V (0.0003051758 V/LSB) 0 V
\@3:SSRMSampleBias: V (0.0003051758 V/LSB) 0 V
\@2:TMFbIgain: V (0.3125000 1/LSB) 1.029125
\@3:TMFbIgain: V (0.3125000 1/LSB) 1.000000
\@2:TMFbPgain: V (0.3125000 1/LSB) 5.089203
\@3:TMFbPgain: V (0.3125000 1/LSB) 0
\@2:DeflLowPassFrequency: V (0.001000000 kHz/LSB) 2.500000 kHz
\@3:DeflLowPassFrequency: V (0.001000000 kHz/LSB) 10.00000 kHz
\@2:FricLowPassFrequency: V (0.001000000 kHz/LSB) 2.000000 kHz
\@3:FricLowPassFrequency: V (0.001000000 kHz/LSB) 10.00000 kHz
\@2:TMDeflLowPassFrequency: V (0.001000000 kHz/LSB) 2.500000 kHz
\@3:TMDeflLowPassFrequency: V (0.001000000 kHz/LSB) 2.000000 kHz
\@2:CurrentLowPassFrequency: V (0.001000000 kHz/LSB) 10.00000 kHz
\@3:CurrentLowPassFrequency: V (0.001000000 kHz/LSB) 10.00000 kHz
\@2:OtherModesLowPassFrequency: V (0.001000000 kHz/LSB) 2.000000 kHz
\@3:OtherModesLowPassFrequency: V (0.001000000 kHz/LSB) 2.000000 kHz
\@2:STMFbIgain: V (0.3125000 1/LSB) 1.000000
\@3:STMFbIgain: V (0.3125000 1/LSB) 1.000000
\@2:STMFbPgain: V (0.3125000 1/LSB) 2.000000
\@3:STMFbPgain: V (0.3125000 1/LSB) 0
\@2:AFMFbIgain: V (0.3125000 1/LSB) 10.00000
\@3:AFMFbIgain: V (0.3125000 1/LSB) 1.000000
\@2:AFMFbPgain: V (0.3125000 1/LSB) 10.00000
\@3:AFMFbPgain: V (0.3125000 1/LSB) 0
\@2:FastZTMFbIgain: V (0.03125000 1/LSB) 1.000000
\@3:FastZTMFbIgain: V (0.03125000 1/LSB) 1.000000
\@2:FastZTMFbPgain: V (0.03125000 1/LSB) 1.000000
\@3:FastZTMFbPgain: V (0.03125000 1/LSB) 0
\@2:InputIgain: V (0.001525879 1/LSB) 10.00000
\@3:InputIgain: V (0.001525879 1/LSB) 10.00000
\@2:InputPgain: V (0.001525879 1/LSB) 10.00000
\@3:InputPgain: V (0.001525879 1/LSB) 10.00000
\@2:VertDeflLowPassFrequency: V (0.001000000 kHz/LSB) 10.00000 kHz
\@3:VertDeflLowPassFrequency: V (0.001000000 kHz/LSB) 10.00000 kHz
\@2:TRLockInBW1: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@3:TRLockInBW1: V (0.001000000 kHz/LSB) 0.5000000 kHz
\@2:Bias Lower Limit: V (0.0003051758 V/LSB) 0 V
\@3:Bias Lower Limit: V (0.0003051758 V/LSB) 0 V
\@2:Bias Upper Limit: V (0.0003051758 V/LSB) 0 V
\@3:Bias Upper Limit: V (0.0003051758 V/LSB) 0 V
\@2:Output 1 Input Data: S [Height] "Height"
\@2:Output 2 Input Data: S [Height] "Height"
\@2:Output 1 Output Data: S [Analog3] "Analog3"
\@2:Output 2 Output Data: S [Analog3] "Analog3"
\*Ciao force list
\Scan rate: 1.0016
\Forward vel.: 67.8543
\Reverse vel.: 67.8543
\Samps/line: 512 512
\Samps/line, retrace: 512
\Ave lines: 1
\Display mode: Both
\Trigger mode: Off
\Rev. trigger mode: Off
\Trig slope: Positive
\Plot start: 0
\Plot end: 1
\Auto start: Disable
\Auto offset: Off
\Start mode: Calibrate
\End mode: Retracted
\User mode: Advanced
\Ramp delay: 0
\Reverse delay: 0
\Indent setpoint: 0.5
\X Rotate: 0
\Scratch length: 100 nm
\Scratch rate: 1
\Scratch angle: 0
\Size linked: Off
\Lift height: 100
\Drop height: 10
\Columns: 5
\Rows: 1
\Column step: 100
\Row step: 100
\Capture: Off
\XY move on surface: Off
\Volume offset: 0
\force/line: 16 0
\Samples / line: 16 0
\Z direction: Retract
\Center plot: Off
\FV scan rate: 0.0607711
\Tip factor: 0
\True resonance: 0
\Deflection Sensitivity Correction: 1.08
\Tip type: 100W
\Unit display: Metric
\PF PGain : 99.8779
\PF ITime : 1.00044
\Linearization: Enabled
\Closed Loop: Enabled
\Polarity: Negative
\Temperature (Celsius): 21
\PSD Bin Width : 7.62939
\Feedback type: None
\Feedback counts: 0
\@Z scan start: V [Sens. Zsens] (0.006713867 V/LSB) 80.98232 V
\@Z scan size: V [Sens. Zsens] (0.006713867 V/LSB) 37.59398 V
\@Z step size: V [Sens. Zsens] (0.006713867 V/LSB) 0 V
\@Sample period: V (0.1000000 us/LSB) 25.00000 us
\@4:Image Data: S [ImageOff] "Off"
\@4:Trig threshold Off: V (1.000000 1/LSB) 0
\@4:Trig rev. threshold Off: V (1.000000 1/LSB) 0
\@4:Threshold step Off: V (1.000000 1/LSB) 0
\@4:Ramp channel: S [Zsweep] "Z"
\@4:Ramp size Zsweep: V [Sens. Zsens] (0.002861097 V/LSB) 33.87288 V
\@4:Ramp offset Zsweep: V [Sens. Zsens] (0.002861097 V/LSB) 51.12194 V
\@4:Ramp Begin Zsweep: V [Sens. Zsens] (0.002861097 V/LSB) 0 V
\@4:Ramp End Zsweep: V [Sens. Zsens] (0.002861097 V/LSB) 0 V
\@4:Feedback value Zsweep: V [Sens. Zsens] (0.002861097 V/LSB) 0 V
\@4:Z display Zsweep: V [Sens. Zsens] (0.002861097 V/LSB) 76.77853 V
\*Ciao force image list
\Data offset: 40960
\Data length: 2048
\Bytes/pixel: 2
\Data type: AFM
\Do zoffder: 0
\Note:
\Plane fit: 0 0 0 0
\Frame direction: Up
\Stage X: 0
\Stage Y: 0
\Stage type: N/A
\Profile length: 0
\Profile speed [um/s]: 0
\Capture start line: 0
\Scan Line: Main
\Realtime Planefit: Line
\Offline Planefit: None
\Force display mode: Normal
\Spring Constant: 1.15455
\Separation Scale: 0
\Force offset: 0
\Z magnew force: 1
\Invert: Normal
\Time channel: 0
\Force TMR: 1.013949e-009
\Scope coupling: DC
\Smoothing filter: 0
\Graph function: i
\STS I range: 15000
\STS DLI/DLV range: 10
\STS DLI/DV range: 10
\STS DI/DV range: 10
\STS Log(i) range: 10
\Ave points: 1
\Force Phase Offset: 0
\Effective BW: 1025.64
\X Data Type: Ramp
\@4:Image Data: S [DeflectionError] "TM Deflection Error"
\@Z magnify: C [4:Z scale] 1.000000
\@4:Z scale: V [Sens. DeflSens] (0.0003750000 V/LSB) 24.57600 V
\@4:Z offset: V [Sens. DeflSens] (0.0003750000 V/LSB) 0 V
\@4:FV scale: V [Sens. DeflSens] (0.0003750000 V/LSB) 24.57600 V
\@4:Volume offset: V [Sens. DeflSens] (0.0003750000 V/LSB) 0 V
\@4:Ramp size: V [Sens. Zsens] (0.002861097 V/LSB) 33.87288 V
\@4:Ramp offset: V [Sens. Zsens] (0.002861097 V/LSB) 51.12194 V
\@4:Ramp Begin: V [Sens. Zsens] (0.002861097 V/LSB) 0 V
\@4:Ramp End: V [Sens. Zsens] (0.002861097 V/LSB) 0 V
\@4:Z display: V [Sens. Zsens] (0.002861097 V/LSB) 76.77853 V
\*File list end