Peak Force Setpoint changing significantly during large scan size imaging - Possible reasons?

rated by 0 users
Not Answered This post has 0 verified answers | 4 Replies | 2 Followers

Top 500 Contributor
3 Posts
Points 33
ksinniah posted on Wed, Jul 11 2012 9:10 AM

Hi All,

I have the following problem with the peak force setpoint changing significantly during a 10-um scan size imaging (i.e changing from ~ 500 pN to about 3-4 nN's), while this problem does not show up when imaging the same sample at smaller scan sizes (i.e. less than 5 um).

The sample contains proteins attached to about ~50 nm particles on a mica surface and are imaged in fluid using scanasyst fluid + and scanasyst fluid probes in a multimode 8 system using a vertical engaging E-scanner. When imaging a 10-um scan the peak force setpoint starts around 500-700 pN's during the image, and when the probe images over a large cluster of particles, the setpoint changes to about 2 nN and then continues to steadily increase during the scanning of the image.  If I reduce the scan size to about 4 um or less, I have a steady peak force setpoint (~ 1 nN) while scanning the image. 

A major concern I have is can I trust the image data with such variations in peak force setpoint?

Thanks in advance for any response.

best,

--Kumar

  • | Post Points: 13

All Replies

Top 10 Contributor
60 Posts
Points 651

After you finish a scan with the setpoint set to automatic, take the same picture with the auto setpoint in off position and set the setpoint first to a low value (60O pN) and take one with the hight setpoint ( 3 nN) and compare the height mesurements of the three pictures. I expect ther will be not mutch difference in the height, but there wil be differences in the modules, adhesion, etc due to different force curves and as a result of that different contact times and deformations.

So I expect it depents on what you want to see.... 

"Meten is weten" as we say in dutch, tranlated it is something like : Measuring is knowing.... Big Smile

Joop

  • | Post Points: 10
Top 10 Contributor
60 Posts
Points 651

After you finish a scan with the setpoint set to automatic, take the same picture with the auto setpoint in off position and set the setpoint first to a low value (60O pN) and take one with the hight setpoint ( 3 nN) and compare the height mesurements of the three pictures. I expect ther will be not mutch difference in the height, but there wil be differences in the modules, adhesion, etc due to different force curves and as a result of that different contact times and deformations.

So I expect it depents on what you want to see.... 

"Meten is weten" as we say in dutch, tranlated it is something like : Measuring is knowing.... Big Smile

Joop

  • | Post Points: 13
Top 500 Contributor
3 Posts
Points 33

Thanks Joop.  I will try that and verify the heights are the same.  Best.  --Kumar

  • | Post Points: 10
Top 500 Contributor
3 Posts
Points 33

As you said, the heights of the particles are not changed when imaging a 10-um size image  (with a calibrated vertical engage E scanner) when the peak force setpoint is set at ~ 1 nN force and then at ~ 3 nN force.  However, the particle height distributions are narrower in the higher peak force setpoint (and on auto mode) compared to the lower fixed setpoint (i.e at 1 nN or less).  The quality of the image too seems low resolution, and there seems to be a "helicoptering effect" observed on top of all tall features (i.e. features around the size of 15-20 nm's). 

We don't see this problem when scanning in air using scanasyst air tips and the problem is much less (hardly noticeable) when scanning in fluid using smaller scan sizes (i.e < 3-um scan sizes).  --Kumar

  • | Post Points: 10