Re-loading PeakForce QNM Calibration values: Is it possible?

rated by 0 users
Answered (Not Verified) This post has 0 verified answers | 1 Reply | 2 Followers

Top 200 Contributor
5 Posts
Points 60
Jerome S. Jourdan posted on Fri, Aug 10 2012 12:40 PM

Is it possible to re-enter values for Deflection Sensitivity, Spring Constant and Tip radius into PeakForce QNM experiments?  After calibration/imaging I may want to switch to TappingMode to image the same area, but this means that when I go back to PeakForce QNM mode the calibration values are gone.  I can re-enter spring constant and tip radius values into the extended menu, but cannot find where to enter deflection sensitivity. I have Ver. 8.15.

Any thoughts? 

  • | Post Points: 12

All Replies

Top 10 Contributor
280 Posts
Points 6,221
Bruker Employee

Sure, you can enter Spring Constant and Tip Radius into the "Cantilever Parameters" parameter list and deflection sensitivity into the Calibrate>Detector dialog (look under the Calibrate menu).  Alternatively, you can uncheck the "Ignore Probe Parameters" box in the experiment selector when you switch from one experiment to another...

--Bede