Remove deflection sensitivity

rated by 0 users
Answered (Verified) This post has 1 verified answer | 2 Replies | 2 Followers

Top 500 Contributor
2 Posts
Points 22
jawarner posted on Fri, Sep 7 2012 12:21 PM

I understand how to calibrate my deflection sensitivity, but how do I UNDO it?  If instead, I would like to view the deflection in mV instead of nm, is there a simple (non-workaround) way to do this?  The reason this is an annoyance is because I adjust my setpoint in mV, but can no longer see my free oscillation amplitude in the RAMP in mV.

Thanks,

Jake

  • | Post Points: 12

Answered (Verified) Verified Answer

Top 10 Contributor
280 Posts
Points 6,221
Bruker Employee

To "undo" the deflection sensitivity calibration just go to Tools>Calibrate>Detector and type 1 "" for the deflection sensitivity (or amplitude sensitivity or amplitude error sensitivity in tapping mode).

Another way is to look for the Units parameter in the scan parm list or ramp parameter list and set it to "Volts". 

--Bede

  • | Post Points: 13

All Replies

Top 10 Contributor
280 Posts
Points 6,221
Bruker Employee

To "undo" the deflection sensitivity calibration just go to Tools>Calibrate>Detector and type 1 "" for the deflection sensitivity (or amplitude sensitivity or amplitude error sensitivity in tapping mode).

Another way is to look for the Units parameter in the scan parm list or ramp parameter list and set it to "Volts". 

--Bede

  • | Post Points: 13
Top 500 Contributor
2 Posts
Points 22

Thanks Bede.

  • | Post Points: 10