Can you switch your setpoint units back from nm to mV (ie. Classic Setpoint Mode) in NS8.15

rated by 0 users
Answered (Verified) This post has 1 verified answer | 1 Reply | 2 Followers

Top 100 Contributor
8 Posts
Points 93
Jakub Holan posted on Mon, Sep 10 2012 9:06 AM

Hello,

the setpoint value for every mode is displayed in nanometers in NanoScope 8.15. Could this be switched back to mV as in the older versions? I know this can be done during installation - ie. you can set that you want the Classic Setpoint Mode. But I think someone mentioned to me that it could also be set later in the software, is this possible?

Thank you

  • | Post Points: 12

Answered (Verified) Verified Answer

Top 10 Contributor
280 Posts
Points 6,221
Bruker Employee
Verified by Jakub Holan

Yes.  Look in the menu under Tools>Options>Classic gain and setpoint mode.  If it is greyed out, clicking on the scan parmlist in scan mode prior to going to the menu should help.

--Bede

  • | Post Points: 11