Strange linear trenches...

rated by 0 users
Not Answered This post has 0 verified answers | 3 Replies | 3 Followers

Top 50 Contributor
23 Posts
Points 275
lgc posted on Mon, Dec 17 2012 7:08 PM

 

When I deal with wide scans I have been encountering these strange linear trenches that compound the noise issues in my images. I've been playing with gains, scan rates and offset voltages, but I haven't been able to eradicate this pest.

I would appreciate the Forum's advice. Thank you and best regards.

 

 

  • | Post Points: 12

All Replies

Top 500 Contributor
3 Posts
Points 34

I think it could be caused by contamination from the sample (or tip).  It should also appear on smaller scan sizes (~50um).

  • | Post Points: 12
Top 50 Contributor
23 Posts
Points 275
lgc replied on Tue, Dec 18 2012 6:46 PM

Would what you mention be the only possible cause? Thanks.

  • | Post Points: 12
Top 25 Contributor
29 Posts
Points 319

This is a very large scan, with a high Z scale as well. You don't say what the sample is, or the Scanning conditions, but my first thought would be to try a very low scan rate, to compensate for the very long scan lines (i.e. Tip velocity = Scansize x scan rate). You could try scan rates of 0.25 Hz, and see if that helps.

Tip and/or sample contamination can also be the cause as mentioned above. 

Also make sure you are using the correct driving frequency (if in tapping mode), by re-adjusting the tune close to the surface.

pete.

  • | Post Points: 10