catalyst: issue with DNP-10 probe

rated by 0 users
Answered (Verified) This post has 1 verified answer | 2 Replies | 1 Follower

Top 200 Contributor
5 Posts
Points 56
CelineHeu posted on Mon, May 6 2013 6:59 PM

Hi,

I am using the BioScope Catalyst.

I can’t manage to engage in fluid with the DNP-10 probe (the “D” cantilever). The motor don’t move, I obtain a false engagement, even if I’m going deep in the liquid (to avoid any false engage due to the liquid surface).

I have not this issue in air.

I also tried to remove the “C” cantilever from the substrate.

Have you ever noticed the same problem? What could I do to fix it?

Best regards

  • | Post Points: 12

Answered (Verified) Verified Answer

Top 50 Contributor
23 Posts
Points 273
Answered (Verified) Janne replied on Tue, May 7 2013 3:29 AM

Go to the Microscope menu and select Engage Settings...

Then lower Engange Gain (e.g. 0.2 or so), and also lower the Engage Step (e.g. to 5um). Maybe this helps. Try different gains.

  • | Post Points: 13

All Replies

Top 50 Contributor
23 Posts
Points 273
Answered (Verified) Janne replied on Tue, May 7 2013 3:29 AM

Go to the Microscope menu and select Engage Settings...

Then lower Engange Gain (e.g. 0.2 or so), and also lower the Engage Step (e.g. to 5um). Maybe this helps. Try different gains.

  • | Post Points: 13
Top 200 Contributor
5 Posts
Points 56

Hi Janne,

Thank you for your useful advice. It worked well.

 

  • | Post Points: 10