problem in e-scanner of nanoscope IIIa

rated by 0 users
Answered (Verified) This post has 1 verified answer | 3 Replies | 2 Followers

Top 75 Contributor
14 Posts
Points 162
Deepak posted on Tue, May 28 2013 9:31 AM

We have been using Nanoscope IIIa instrument which has recently developed a problem. The e-scanner is not able to image the samples and the features appear to be stretched in one direction. I am attaching two images of the same sample, one taken with the e-scanner and  another with the J-scanner. The stretching in E-scanner is evident.  Any suggestions or help will be greatly appreciated.

 

 

 

another another from the j-scanner.  

  • | Post Points: 12

Answered (Verified) Verified Answer

Top 10 Contributor
60 Posts
Points 651
Verified by Deepak

My first guess is that there is something wrong with your calibration file of the e-scanner, assuming you select the right calibration file. If you still have the previous calibration file from before the last calibration try loading that one and try again. The fastY sensitivity looks wrong ( top of the picture is much more stretched than the bottom), or try re-calibrating with the grid

  • | Post Points: 14

All Replies

Top 10 Contributor
60 Posts
Points 651
Verified by Deepak

My first guess is that there is something wrong with your calibration file of the e-scanner, assuming you select the right calibration file. If you still have the previous calibration file from before the last calibration try loading that one and try again. The fastY sensitivity looks wrong ( top of the picture is much more stretched than the bottom), or try re-calibrating with the grid

  • | Post Points: 14
Top 75 Contributor
14 Posts
Points 162
Deepak replied on Wed, May 29 2013 10:14 AM

I have tried using the previous calibration file, but the problem persists. Re-calibration is difficult as I am not able to image the grid.

Is it possible that one of the piezo of the scanner has gone bad and is there some way to check this.

 

  • | Post Points: 10
Top 75 Contributor
14 Posts
Points 162
Deepak replied on Fri, May 31 2013 8:30 AM

I have tried using the previous calibration file, but the problem persists. Re-calibration is difficult as I am not able to image the grid.

Is it possible that one of the piezo of the scanner has gone bad and is there some way to check this.

  • | Post Points: 10