CelineHeu posted on Wed, Jul 24 2013 6:13 PM

Hi All,

According to the manual the MM8 is not designed to realise quantitative LFM. does anybody have an experience to share about manual calibration techniques?

regards

Celine

  • | Post Points: 12

All Replies

Top 50 Contributor
20 Posts
Points 220

I'm afraid it's difficult to do with any AFM.

A good and relatively recent review can be found here:

http://iopscience.iop.org/0022-3727/43/6/063001

Good luck,

Nic

 

  • | Post Points: 12
Top 200 Contributor
5 Posts
Points 56

thanks for your answer Nic,

 

I was wondering if you could help me with the choice of a standard sample. I found this one.

Normal 0 false false false EN-AU X-NONE X-NONE MicrosoftInternetExplorer4 http://www.ntmdt-tips.com/products/view/tgx1

 

Celine

 

  • | Post Points: 12
Top 50 Contributor
20 Posts
Points 220

Hi Celine,

the sample you linked to is usually used for scanner calibration, as opposed to cantilever calibration, so I don't think it would be a good choice.

The standard that is typically used for the Varenberg wedge method  is the TGF11 grating from Mikromasch. I can't see it on the Mikromasch website anymore, but it appears to be available here: http://www.nanotoolstore.com/afm_accessories_samples_mikromasch.htm 

Note that this grid is not compatible with all tip shapes - it can give artifacts with low aspect ratio tips, such as those found on many silicon nitride cantilevers.

Nic

  • | Post Points: 10