孙博士,请问roughness 分析时,SSK、SKu、S10z[nm]各是什么含义?它们可以表征image的信噪比吗?
Ssk称为偏度(Skewness),它描述了数据分布偏离对称的程度。当分布左右对称时,偏度为0。当偏度大于0时,该分布为右偏,即表面上有较多高于基线平面的高峰数据。当偏度系数0时,该分布左偏,即表面有较多低于基线平面的低谷数据。其定义为图片上每一个像素点的高度数据的三次方加和,再除以像素点的个数,再除以均方根粗糙度的三次方。该值的绝对值越大,说明数据偏离中心的程度也就越大。
Sku称为峰度(Kurtosis),它是用来反映分布曲线顶端尖峭或扁平程度的指标。当为正态分布的时候,峰度为3(有的地方会减掉这个3让它等于0)。当峰度大于0时,数据的分布比较平,峰度小于0时,数的分布比较尖。其定义为图片上每一个像素点的高度数据的四次方加和,再除以像素点的个数,再除以均方根粗糙度的四次方。
我们离线软件中的粗糙度参数是以Skewness、Kurtosis表示Ssk和Sku,并没有S10z,请问您是在哪个界面下看到这个参数,这样可以找到相应的名称解释。
我们这里只有这样一个关于S10z的解释:S10z[nm] - the average height of the five highest local maximums plus the average height of the five lowest local minimums。
那究竟这两个参数是否可以表征图片的信噪比呢?或者说是不是Ssk越偏离0,Sku越偏离3,就说明图片受噪音的影响就越大?
所以S10z在您的软件中的意思是五个最高点的平均值加上五个最低点的平均值。在我们的新的NSA离线软件中没有这个参数的相应参数。请问您使用的软件是?
这些参数均与噪声没有直接关系。如果要评估噪声,一般是使用均方根粗糙度Rq。具体方法是扫描一个0nm的位置,进行一阶拉平后将Rq与仪器噪声指标进行对比。
假定你的图片中有噪声,同没有噪声相比,Ssk并不能说明问题,因为两个图都可能是对称分布的。噪声也有正有负。Sku则能部分反映,因有噪声时数据的分布范围比较宽,但一般仪器的指标不会用它来衡量。S10z类似Ssk。所以最直接的比较还是均方根粗糙度。
贴一个仪器指标的一般测试方法(以Icon原子力为例):
仪器指标检测:
扫描管范围检测:XY:____ m,应不小于85m______(√);Z:____ m,应不小于9.5m______(√)
测试方法:在XYZ开环下,将XYZ方向的扫描电压设置为最大值,检查最大扫描电压对应的扫描范围。
系统噪音检测: ____ nm RMS ,应小于0.03nm______(√)
测试方法:打开XY方向的闭环,在tapping模式下扫描0 nm范围,扫描速度2.44 Hz获取Height图像,测量图像的Rq值。
扫描管Z方向闭环噪音水平: ____ nm RMS,应小于0.035nm ______(√)
测试方法:打开XY方向闭环,假进针,扫描范围0.01 nm,扫描速度2.44 Hz,获取Height sensor图像,测量图像的Rq值。
X方向的精确度: <2% ____(√)
测试方法:扫描标准样品,确保测量的尺寸与标准尺寸的偏差在1%以内。
Y方向的精确度: <2% _____(√)
Z方向的精确度: <2% _____(√)
闭环扫描管的XY方向正交性: ____°(90°±2°)
测试方法:扫描标准样品,测量所获得的光栅图像X边缘与Y边缘的角度,确保此角度在90°±2°范围内。
样品台XY方向移动范围: X方向 _____mm,应≧180mm _____(√)
Y方向 _____mm,应≧150mm _____(√)
好的,我们去试试。谢谢孙博士!
孙老师解释的很清楚。我也是第一次看到中文的翻译。受教了。
其实对各个参数,Nanoscope Analysis的说明书上也有介绍。在使用Nanoscope Analysis的时候按F1,说明书就出来了。一下是说明书来的,粘贴过程公式丢失,详见说明书。
如果你需要最新版的Nanoscope Analysis的软件,我可以提供下载链接。
Statistics used by the Roughness routine are defined in this section. The terms are listed alphabetically. Most are derived from ASME B46.12 (“Surface Texture: Surface Roughness, Waviness and Lay”) available from the American Society of Mechanical Engineers.
Table 3: Roughness Results
谢谢Teddy的补充!
非常感谢!还望提供链接。谢谢!
留个邮箱或者发request给我到 teddy.huang@bruker.com
谢谢
孙博士,我用两套参数测试roughness(tapping mode)。 其中Amplitude较高的得到的Rq=0.114nm SSK=0.0571 SKU=3.06
Amplitude较低的得到的Rq=0.14nm SSK=0.116 SKU=3.3
然后我有用scan size=0um 去测试Rq, 其中Amplitude较高的得到的Rq=0.0694nm 其中Amplitude较低的得到的Rq=0.0534nm
那么这两套参数谁的信噪比更好呢?怎么解释这种现象?谢谢!
Tapping mode下测roughness和很多因素相关,必须固定探针以及探针和样品间的作用力来测量,这样得到的结果才能进行比较。若在不同的Amplitude下,由于探针与样品的相互作用不同,且因响应时间不同而导致Gain的最优设置不同,都会对Roughness有影响。所以不能直接比较。
请问您的应用是什么,为什么要比较信噪比?
我们测试的roughness小于0.4nm的样品。Amplitude小的我们用的gain值也会相应的减小。
请问噪音到了什么程度就可能影响到Rq的结果呢?有没有一个衡量的标准?
如果使用Low Force Tapping来测超平样品,在Amplitude减小时由于响应时间随之减小,相应的Gain应该增大,否则就失去了小振幅的优势。请参考以下Low Force Tapping的步骤和注意事项:
(以MM8为例,软件版本为8.1x,其他版本软件原理类似)
3. High resolution imaging with tapping Mode
A. Probe selection: OLTESPA or FESPA probes are suitable for high resolution imaging under tapping mode. Stiffer probe like TESPA causes too much impact to the sample, it is not recommended due to the worse force control. Probe without coating such as TESP, RTESP is also not recommended, because the signal-noise ratio becomes lower due to low SUM signal.
B. Sample: In this case we use alkane sample for high resolution imaging experiments.
C: Force control: Tapping force control is more difficult than Peak Force Tapping force control, because the relationship between tapping amplitude and tapping force is more complex. Under tapping mode, to control force, one should finely adjust amplitude setpoint and drive amplitude. The relationship between tapping amplitude and tapping force was demonstrated by Chanmin Su, Lin Huang, et al. (2003) and Shuiqing Hu (2007). This is shown in following picture. Tip wear often happens at the middle amplitude setpoint values. To achieve high resolution in tapping mode, one can use “Low force tapping” technique invented by Lin Huang et al. which is operating at very low amplitude setpoint and produces very low tip-sample interactions. The low setpoint operation enables high feedback loop gain, which improves tracking and reduces tip wear. However, sometimes it can results in 0 amplitude on rough samples at high scan rate.
D. D. Noise control: For MultiMode SPM system, put the microscope on the triple to reduce the environmental noise.
E. Procedure:
(1) Start the NanoScope v8 software by double clicking the NanoScope software icon on the Windows desktop.
(2) In “Select Experiment” Window, choose “Tapping Mode” and correct experiment group due to your experiment, and then click “Load Experiment”.
(3) Go to “Microscope” Menu, select “Engage Settings”. “Engage Parameters” pops up, in this window, set “Sew tip” to “Yes”, which can protect probe during engage process. Then click ‘OK”.
(4) Click “Setup” icon on the workflow toolbar, in “Tune cantilever” windows, click “Manual Tune” to open “Cantilever Tune” window. On “Auto Tune” Panel, modify the default “Target amplitude” to a smaller value, usually “100 mV”, set “Peak offset” to “5%”, click “Auto Tune” button. Adjust “Data Scale” in “Channel 1” Panel and “Sweep Width” in “Graph” Panel to make Amplitude-Frequency curve easy to see. Record the “Drive Amplitude” value. Click “Exit” to close “Cantilever Tune” window.
(5) Set initial parameters. Click “Expended Mode” to display more parameters. In “Scan” Panel, set “Scan Size” to “0.00 nm”, “X Offset” and “Y Offset” to “0.000 nm”, “Scan Rate” to “0.977 Hz”. In “Feedback” Panel, set “Integral Gain” to “0.300” and “Proportional Gain” to “1.000”, set “Lock-in BW” to “5k Hz”. In “Limits” Panel, set “Amplitude Limit” to “1000 mV” and “Z Limit” to its max range. Set Data Channel of Height, Amplitude Error, etc.
(6) Explore sample surface under OMV video, the best place to be imaged is flat area around some black spots. Select Microscope > Engage or click the Engage icon on the workflow toolbar to engage the tip.
(7) Reduce the “Drive Amplitude” by 2/3 so that the free amplitude in air should be 60~70 mV now if the “Target Amplitude” is set to “100 mV” when tuning cantilever. Reduce the “Amplitude Setpoint” to 12~15 mV. Increase “Integral Gain”, and monitor the “Amplitude Error” signal. Slightly decrease “Integral Gain” if oscillation is observed. The peak value of “Amplitude Error” should be less than 10 mV. In “Limits” Panel, decrease “Z Limit” to “1.000 um” to increase vertical resolution. Capture the image.
(8) After imaging, select Microscope > Withdraw or click the Withdraw icon on the workflow toolbar to withdraw the tip.
请指明您的仪器型号,如果是新型号的仪器做这类应用请使用ScanAsyst模式。