请问在Dimension Icon在模式切换时出现:“Error log: sensor pre-amp-offset of z center(0.6262) is over the 0.6 limit!”

rated by 0 users
Answered (Verified) This post has 1 verified answer | 6 Replies | 2 Followers

Top 150 Contributor
8 Posts
Points 94
zengqibin posted on Tue, Mar 31 2015 9:00 PM

孙博士在吗?我们的Dimension Icon这两天在模式切换时会出现:“Error: sensor pre-amp-offset of z center(0.6262) is over the 0.6 limit!”,但扫图又基本正常,我想请问您这是什么地方有问题吗?谢谢您!

  • | Post Points: 12

Answered (Verified) Verified Answer

Top 10 Contributor
363 Posts
Points 4,300
Bruker Employee
Verified by Hao Sun

工程师已经确认为高压板的问题。

  • | Post Points: 13

All Replies

Top 10 Contributor
363 Posts
Points 4,300
Bruker Employee

闭环子系统可能引起此问题。此问题涉及到较复杂的诊断,下周将有工程师联系上门诊断。

  • | Post Points: 12
Top 150 Contributor
8 Posts
Points 94

非常感谢您的答复,请问我们现阶段是否应停止使用这台仪器?

  • | Post Points: 12
Top 10 Contributor
363 Posts
Points 4,300
Bruker Employee

闭环子系统相对比较独立,并且如果是为了做形貌成像,Z向即使Height Sensor得不到正确数据,也可以先参考校准过的Height通道数据。在工程师到之前您可以继续使用。

  • | Post Points: 12
Top 150 Contributor
8 Posts
Points 94

好的,谢谢您!

  • | Post Points: 12
Top 10 Contributor
363 Posts
Points 4,300
Bruker Employee
Verified by Hao Sun

工程师已经确认为高压板的问题。

  • | Post Points: 13
Top 150 Contributor
8 Posts
Points 94

是的,主要问题是z+ piezo的高压放大芯片损坏,现在仪器已基本恢复正常,感谢您及几位工程师的热心帮助,谢谢!

  • | Post Points: 10