Hello.

I am working with a Bioscope Catalyst with a Nanoscope V Controller.

At the start of the vertical ramp when the tip has been pulled up into air, I have recorded a 0.6V~1.3V Vertical Deflection signal on the E-Box instead of the expected 0.00V. This caused the resulting vertical ramp to have a baseline offset and not start out at 0nm deflection.

I have made sure to center both the vertical and horizontal deflection values to as close at 0.00V before engaging.

I know this question has been asked before and that the baseline offset can be corrected using the analysis software, but I still don't quite understand what is causing this offset in the vertical deflection value?

The explanation provided in the analysis software manual indicated that the baseline offset was caused by artifacts but did not expand into further detail. 

Thank you!
-Yen  

 

  • | Post Points: 12

All Replies

Top 10 Contributor
62 Posts
Points 660

Hi Yen,

May I have your email? I'll get our Bioscope Catalyst expert to help you out. You can also contact me at Teddy.Huang@bruker.com

Thanks,

Teddy

  • | Post Points: 12
Top 500 Contributor
4 Posts
Points 51

Hi Teddy,

My email is yenhseu@gmail.com

Thank you!
Yen  

  • | Post Points: 12
Top 10 Contributor
62 Posts
Points 660

Hi Yen,

I've forward your question to our expert and you're copied.

Please expect delay in response as he is on an internal project until later this week.

Thanks,

Teddy