Ampltiude/Phase distance curves in tapping mode

rated by 0 users
Not Answered This post has 0 verified answers | 1 Reply | 1 Follower

Top 500 Contributor
3 Posts
Points 34
msouier posted on Tue, Oct 13 2015 2:55 AM

I have a very basic question since I am new in using Multimode 8 AFM.

I would like to perform Amplitude/Phase-Distance curve. I Failed to find the steps in the manual,

Can you please show me the steps to record the curve in tapping mode,

Thanks & regards,

Souier.

  • | Post Points: 12

All Replies

Top 25 Contributor
37 Posts
Points 422
Bruker Employee

Hi Souier,

 

This should be very simple. After you have engaged the probe on the sample surface in the "workflow" bar on the left of the screen the next node/ icon after engage should be "ramp". Click on this and it will take you to the force curve GUI which gives you three channels and a bunch of parameters on the left. Simply assign the channels as you wish i.e. Ch1: TM deflection versus distance Ch2: Amp versus distance Ch3 Phase versus distance. In the ramp parms on the left remember to set a trigger. I would recommend an absolute trigger on amplitude, trigger direction negative (as the amplitude will decrease) so that the probe does not overshoot and damage the tip/ cantilever/ sample as it touches the surface.

 

Best regards

  • | Post Points: 10