I cannot view your images. Did you try to upload them using "My Files" (bottom right of your profile homepage) and to include it using "Insert Media"?
Dietmar
This is the first image, calibration in fluid
And the second, calibration in air
Thank you!
Before you do the fit you should zoom in a lot more on the peak so that your x-axis is only ~10 to 30 eV. Than your baseline is "straight"and the fit will be good because with the axis like this you will get very wrong fittings and as a result wrong spring constants. You zoom in with "control" mouse click, but I think you will know that
Please let me know if you managed to solve your problem.
We were able to zoom in and fit the curve more closely but spring constant still measures a high spring constant. It seems we will need a reference probe in order to correctly calibrate this probe.
I guess the spring constant was specified to the original cantilever and since you have extra mass attached, the resonance frequency will shift and back calculation of spring constant from thermal tune will be offseted.
Ang Li