孙工你好,
我上次发了求助信,
http://nanoscaleworld.bruker-axs.com/nanoscaleworld/forums/t/909.aspx
后来跟换电池后问题解决了。
但我使用的时候发现原来 Illumination 是计算机控制的,软件中有调节 Illumination 的选项,现在没有了,只能手动调节 Illumination 的旋钮。
Event中有这个error
Event ID: 11
Source: Serial
The hardware resources for COM3 are already in use by another device.
Event ID: 24
A conflict has been detected between two drivers which claimed equivalent IRQs. Driver Serial, with device <\Device\Serial10000.Translated>, claimed an interrupt with Level in data address 0x28, vector in data address 0x2c and Affinity in data address 0x30.
是哪里没有设置好吗?技术支持说可能软件中有设置,可是软件中没有找到啊?
不知道你是否了解这些端口的设置。
先谢谢了。
启动软件时,选择Microscope Select,打开显微镜选择菜单,上头有个serial设置,可以更改COM口的配置。正常的配置是COM1是trackball,COM2是stage,COM3是Illumination。若更改了实际的串口配置,就要在这里更改过来。如果某个串口有问题,也可以将有问题的设备连接到其他串口并在这里更改。
孙博你好,谢谢回复。我找到Microscope Select菜单以及serial窗口了,也看到了COM1是trackball,COM2是stage,但没有COM3,也没有地方可以增加一个端口。不知道怎么加?
另外不知道是不是跟计算机本身的端口设置有关系。计算机控制面板的端口设置中,COM3的 I/O Address 和 IRQ 有固定需要的设置吗?因为当时以为COM3有问题,把COM3删掉了,不记得原来是什么设置了。
谢谢。
如果电脑上的COM3和COM4是用的扩展卡,重新插一下扩展卡,然后检查一下BIOS设定。BIOS中不要禁用COM3和COM4. BIOS的I/O和IRQ的设定见附件。
孙工你好:
我公司有一台Dimension 3100, 原来通电后可看到Dimension 3100 controller上的AIR PRESSURE有压力,现在没有压力了,但主电脑关机时,压力表指针会快速的动一下。
请问这里的压缩空气是用于哪里的;电脑软件运行没有错误提示,为什么?