Hi Dietmar, Are most of the curves in the HSDC capture like this? Or is this an unusual curve? If you look at the force vs. Z plot in the force monitor (in realtime) do you see the same behavior? Or do they look correct in realtime? What version of Nanoscope and Nanoscope analysis are you using? --Bede
-
Hi Fei, The 500KHz HSDC uses an ADC that samples at 500,000 samples/second. There is a hardware antialiasing filter on that ADC at 200KHz. On top of that, some data types have additional low pass filters or lower filters (X, Y sensors for example). The 6.25MHz HSDC uses an ADC that samples at 50,000...
-
Palli, I uploaded your file here: batchHSDC.zip
-
Hi Cosmin, There are separate sections in the header for each data channel. Each of the sections starts with a line "\*Ciao HSDC list". You will find different Data offsets, lengths and scales for each channel. I looked around and found an Igor procedure to read HSDC files! Here is a description...
-
HSDC files are usually so large that they are unmanageable when exported to ASCII. It is usually much more efficient to directly read the native NanoScope files instead. This is especially true if you plan on analyzing a large number of files. Nanoscope files are composed of an ASCII header that contains...
-
Normal 0 false false false Hi All, I am trying to figure out a way to batch export High Speed Data Capture (hsdc) files generated with Veeco’s Nanoscope software. Unfortunately the Nanoscope software does not support batch export for those files. Does anybody know a software that can open and batch...