  • Re: PFM

    Certainly! There is no special requirement to do PFM (just use contact mode and configure the lockins through generic lockin). I recommend putting the AC bias to the tip and reducing the 'amplitude limit' for the lockin that you are using. I think lockins 2 and 3 are preferred depending on your...
    Posted to SPM Digest (Forum) by Bede Pittenger on Thu, Jul 20 2017