  • Re: Debris or artifact?

    That definitely looks like debris is being picked up by the tip, causing a step. Additionally, it looks like you have used a second or third order flatten or a realtime "Bow" fit, causing the steps to turn into arched scan lines. Please take a look at the offline manual section on flattening...