Dear All,

Does anyone know why the DMT modulus image have negative value (no flatten)?

Thanks,

Iwei

  • | Post Points: 12

Answered (Verified) Verified Answer

Top 10 Contributor
129 Posts
Points 1,429
Answered (Verified) Ang Li replied on Sun, Mar 18 2012 8:44 PM

Most likely you have a too high deflection sensitivity, you can try reducing it and modulus image would back to normal.

Ang Li

  • | Post Points: 13
Top 10 Contributor
280 Posts
Points 6,221
Bruker Employee

As Ang Li says, the problem is that your deflection sensitivity is too high.  This causes the force vs. separation plot to bend back so that it appears that the sample deformation decreases with increasing force.  When the DMT model is applied to this data it results in a negative modulus. 

If you want to work with a qualitative measurement, you can just decrease the number (look in the 'Calibrate>Detector...' menu for the 'Deflection Sensitivity').  If you want accurate numbers you need to calibrate the deflection sensitivity in ramp mode on a sample that is clean and very hard (saphire, silicon, etc depending on how stiff your probe is).

--Bede

  • | Post Points: 13

All Replies

Top 10 Contributor
129 Posts
Points 1,429
Answered (Verified) Ang Li replied on Sun, Mar 18 2012 8:44 PM

Most likely you have a too high deflection sensitivity, you can try reducing it and modulus image would back to normal.

Ang Li

  • | Post Points: 13
Top 10 Contributor
280 Posts
Points 6,221
Bruker Employee

As Ang Li says, the problem is that your deflection sensitivity is too high.  This causes the force vs. separation plot to bend back so that it appears that the sample deformation decreases with increasing force.  When the DMT model is applied to this data it results in a negative modulus. 

If you want to work with a qualitative measurement, you can just decrease the number (look in the 'Calibrate>Detector...' menu for the 'Deflection Sensitivity').  If you want accurate numbers you need to calibrate the deflection sensitivity in ramp mode on a sample that is clean and very hard (saphire, silicon, etc depending on how stiff your probe is).

--Bede