Forums
Bruker Media
Community Media
Bruker AFM Probes
SPM Digest
Application Notes
NanoTheater
Website
中文
Brochures & Datasheets
Publications
Probes Catalog
Events
Manuals & Documentation
Presentations
Guide to AFM Modes
News
Journal Club
Webinars & Video
Nanovations
Other
Hi friends,
1) The noise during the CAFM measurement suddenly changed. Before, the current noise was around 2pA (High current sensor 1nA/V). Now it is 10pA. Does it has something to do with the wear of the tip (SCM-PIC)?
2) By the way, I test the current signal on the sample chuck with a new tip. But the current signal appears until the DC sample bias reaches -4.2V or +4.8V. The bias is correctly controlled by the software because I verify this by a multimeter. So the question is why the conductive sample chuck has a threshold voltage? I expected the current image would increase as the bias increase.
Thanks a lot!
Hi Solong,
1. How do you quantify the CAFM noise? It is peak-to-peak or RMS?
2. A couple of things to be aware of:
Chunzeng
Hello Chunzeng,
Thanks a lot for your reply!
1. I quantify the CAFM noise from peak to peak.
2. The Pt/Ir coating is only 20 nm thickness. On the bottom, the tip is also coated with 3nm Cr. This coating is being used during the measurement. I almost cannot get the repeatable CAFM results of one sample. I don´t know when the Pt/ir coating has been totally worn off, when the Cr coating or Si takes a part in the current measurement.