I have found that the image quality of adsorbates on mica (e.g. DNA molecules) depends on the purity of the water used to prepare solutions and to rinse the mica disks.
I like to use water from a local Milli-Q system, but sometimes the water quality is not good enough. Can anyone suggest a commercial source of water?
We have used quite successfully Ultrapure MilliQ water for imaging ofsingle molecules. If the filter cartridge/membrane were not changed asrecommended, there is a good chance that water gets contaminated. We donot change those as frequently as the manufacturer recommends (tooexpensive), and therefore, we have to monitor quality of this water. We found three quite simple tests: 1. pH of good water should be around 7. If it is below ~6 or above ~7.5,there is a good chance that it is contaminated. 2. Light scattering: should be none. The device to check it is verysimple, and can be assembled within an hour. It is a cuvette forfluorescence, which you fill with water to analyze. Put it underanoptical microscope with an objective with working distance >~1mm (>than wall thickness of the cuvette) . Then you illuminate water in yourcuvette from the side with a laser (a good laser pointer might besufficient). If there are any particulate contamination there, you willsee it (starting from the size of a few nm). (P.S. Definitely, oneshould be caution about using lasers, and not trying to send it directlyto your eyes.)3. (probably the most accurate), just look at Raman spectra (a regularfluorescent spectrometer should work). Contaminated water has specificsignatures that are very easy to recognize.
Yes, the quality of miiliQ water can be rather low for low height, dry samples like DNA. Note that it depends a lot on how well the miiliQ machine is maintained! I've compared the quality of various waters, for this aplication, and the best is Sigma
"Water for Molecular Biology"
It's around 40 Euros a litre.
This is mentioned at http://www.fc.up.pt/pessoas/peter.eaton/afm_faq.html#10
We have had the exact same problem in our lab; that is, imaging DNA on mica in air,
and finding particulates on the surface. We spent a long time trying to get to the bottom of it,
and to the best of our knowledge the impurities actually come from two different sources.
When making control samples incubated and washed with water from various sources (MilliQ,
distilled, tap etc.) we noticed a variation between and within sources, and found that BPC Grade
water from Sigma (W3513 - Biotechnology Performance Certified) gave the cleanest and most
reproducible results. The other source of impurities we found was the nitrogen stream used to
dry the samples after washing. What exactly these particulates are we never found out, but by
running the nitrogen through a simple, home-made bubble trap, they largely went away.