Corey Thompson posted on Wed, Jan 29 2014 2:12 PM

I am using a Dimension Icon with NanoScope v8.15 to image multiple locations on very similar samples. In ScanAsyst mode (with auto setpoint and auto z limit set to OFF) when I withdraw to move the stage and then reengage, the peak force setpoint and z range reset to default values. Is there any way to lock these values during a withdraw/reengage cycle?

Thanks very much for your help,

Corey Thompson

  • | Post Points: 14

All Replies

Top 150 Contributor
7 Posts
Points 82
jimmy05 replied on Wed, Mar 12 2014 10:01 AM

Hello Corey

Unfortunatelly I don't know any lock, but you may try to find out your values and when you withdraw change the values and that reengage. The next cycle should hold your values.

Jimmy

  • | Post Points: 10
Top 10 Contributor
129 Posts
Points 1,429
Ang Li replied on Wed, Mar 19 2014 4:11 AM

Hi, Please try to change the parameters before you engage and they should be firstly automatically set upon engage (for a few seconds at the beginning of imaging) and later change back to user preset value during imaging. Please let me know if the procedure doesn't work and I can pass you another procedure to 'lock' the values.