Hi Everyone!
I am an user of an AFM Dimension Icon (Bruker) model.
I would like to measure the coefficient d33 of a polymer thin film doped with piezoelectric nanoparticles. I tried to make some scans in the PFM modality, obtaining the amplitude and the phase data. But I don't know an efficient way to calculate the values in pm/V from the scans.
Is it possible to extract the d33 coefficient from the scans, or are there other easier "tricks"? For example, other type of analysis (ramp)?
Thank you very much for your attention and availability.
Lorenzo
Hi Lorenzo,
To measure the d33 coefficient from the PFM amplitude channel you need to perform the following calculation:
{the value you wish to use from the PFM amplitude channel [V] multiplied by the deflection sensitivity of the cantilever you are using [nm/V]} divided by {16 x the applied bias you are using [V]}
You can see by following the units this will leave you with nm/V
The number 16 is introduced as this is a gain factor used in the instrument.
This calculation is in the manual or help file I believe under the "background contribution" section.
Best regards
Dear Ian,
thank you for your explanation.I saw in the manual the formula that you have cited, and this is a good start point.but I have other questions...1) the deflection sensitivity is a given value of each tip, or it should be calculated? Because I am using SCM-PIT probes, and I have just their deflection sensitivity value (eg 112 nm/V). Instead, in some studies that value was evaluated with a force-displacement curve.2) Regarding "the value you wish to use from the PFM amplitude channel", it is represented by an average value obtained by a single scan, or it is referred to some particular points of a scan? In my case it is a delicate point, since I am studying nanocomposite, so I don't know if it is correct to calculate an average value of the total scanned area or not.
Thank you for your availability,
1) the 112nm/V deflection sensitivity value is nominal, more accurate would be to evaluate it by doing a force curve and measuring the gradient in the contact portion of the curve.
2) this is your choice based on the sample - if your sample was completely homogeneous across the scanned area then you could use the average value of the entire scanned area. If you are interested in certain regions within the scan area then you will need to base the calculation on the amplitude in that particular region. You could gather this value by picking a point, looking at a cross section or small area and taking the average value.
Ok!
thank you Ian for your further explanations.
Best regards,