Dear all:
Now I am using scanning tunneling microscopy and added lock-in amplifier to get STS curve. I want to reduce noise and get dI/dV curve. I got some help from expert who had used STS about connection of lock-in amplified with STM system. I followed him as shown in the picture below, but my STM system does not work properly. I have some problem to operate STM after connection. The software (Nanoscope) always says that tip was engaged from the begining of experiment even the tip was not moved at all. When I turn the toggle switch (from Signal Access Module III, SAM III) of the 'bias' from left (input) to the right (output), the tip moves properly.
My questions are;1) Is the connection good to use STM and STS? And can I get a dI/dV curve with this connection?2) Why it always fails to engage the tip with toggle switch 'input'?3) Which software should I use to get modified curve, Nonoscope or Acquire?4) If I should use Acquire, which parameter I should choose to get dI/dV curve? X axis: I need bias voltage, but software has (Osc Freq, Osc Amplitude, DAC 1-4 voltage, and Digital port) Y axis: The need current depend on bias. But there are no current (amp).
Thank you very very much.Regards,Ki-Rak Kim
Dear Ki-Rak,
Can you send me the manual for the Lock-in Amplifier you are using, I'd like to know what exactly are ADC1, ADC2, A, OSC Out.
Here is my guess:
If the above is right, then you need to change your connection as follows:
When you run I-V, one channel would still be Current vs Bias (I-V); the 2nd channel AuxA vs Bias (dI/dV - V).
The other thing you may try is: with the modified connection, toggle to "Bias Input" only after engage and when you are running I-V.
Try the above, and we go from there.
Best regards
Chunzeng
I read the manual of the lock-in amplifier (model Recovery 7265), my guesses about the signals are incorrect in my previous response, so please disregard it.
The connection diagrams you proposed is correct, albeit may be incompete. It is possible to set it up for either or both Nanoscope or Aquire to read and display I~V and dI/dV~V curves. Lets focus on the former. The first and foremost is to make sure tip will engage the sample surface.
Connections between the Lock-in and SAMIII, based on your diagram:
Settings on the Lock-in side:
Toggle switch positions on SAMIII:
Question and suggestions about your "Adder"
Signal check / diagnosis: use a Oscilloscope to
Our goal for this first step is to make sure tip will engage. We will get to data acqusition and display in the follow-up communication.
Thanks
Thank you very much, Dr. Li.
Your reply and phone call helped me a lot. I follwed your advice and finish to set it up. Now I can engage the tip properly.
But dI/dV curve still has a lot of noise. And there's no difference with and without lock-in amplifier.
I am going to talk to another technician from Veeco today. I hope I can solve every problems today.
Thank you again.
Ki-Rak Kim
Thanks for the update. I'd like to recap some points discussed during the phone call:
To reduce the nosie on dI/dV, you can use a bigger Oscillation amplitude. Note, if the I-V is noisy, then dI/dV will be too.
Now for some people who are famous in some field such as the stars when they go out with their friends or other people. They are often being taken photos secretly by some journalists and they will only realize it until they see it on the newspaper or on the TV. Yes for the stars sometimes they do not mind so, as it is also a way to raise the popularity of them, but sometimes they also mind. However for most people as the normal working people and so on, for them being taken photos by the journalists or recorded down the things that they have said is not their willing and may even have band effect on their life. But if they are trapped in such condition now they can pick up the wireless jammer to help them.
Dear Chunzeng,
I have the same problem about connecting like Ki-Rak.
below is my Instrument and connect:
For lock-in Amplifier setting:
Oscillator frequency to be 3000Hz, and Amplitude 20mV
For STS display:
Channel1: current (I~V)
channel2: AuxA (dI/dV~V ?(I'm not sure...))
My questions are:
Thank you a lot!Regards,
Hung-Jen