Apologies if these questions have been asked ad-nauseum, but I've been unable to find them in my searches.
I'm performing Peak Force QNM in air on polymer samples (ABS) and have a couple of questions regarding calibration (Nanoscope V and MultiMode 8):
1. During Thermal Tune I get a status message indicating "Internal Thermal Tune Gains have not been calibrated. Contact Service". I'm able to acquire a thermal tune and get a spring constant, but does this error message indicate wrong values for 'k'? A latest calibration of a ScanAsyst-Air probe gave me 0.18 N/m value.
2. When performing tip qualification on the Titanium roughness sample which value should I choose for Estimated Tip radius? ETR 1 (smaller) or ETR 2 (larger ) value? I seem to to get closer values for Reduced Modulus when I scan the PS/LDPE sample (PS ~ 3 GPa).
3. Lastly, If I'm performing a Peak Force scan on a polymer surface and see nice images for Height, Peak Force Error, Amplitude and Dissipation, but get lousy contrast (lots of streaking) in the Modulus and Log Modulus images, is this a function of wrong probe?
Thanks,
Jerry
Hi Jerome,
I hope that helps!
--Bede