KAU_ECE posted on Tue, Jan 14 2014 10:51 AM

Hi all

 

We are trying to measure the DMT modulus using PeakForce QNM for some nano-fibers but for some reason we are seeing negative values. 

While we are taking the images, the data shown on the screen shows positive values ranging from 50-100 MPa for the fibers (see the first screenshot below), but when we capture the complete image and try to measure the values again, it shows negative values.

 

 

 

 

Why does the fiber section show negative values in the above image? It seems like Nanoscope is scaling the values somewhere. How can we retain the original values as it shows while imaging.

 

 

Thank you

  • | Post Points: 14

Answered (Verified) Verified Answer

Top 25 Contributor
29 Posts
Points 341
Bruker Employee
Verified by KAU_ECE

When you collected your data, the Offline Planefit parameter (called "OL Planefit" in the software) was set to "Offset."  This will offset the average Z value of the image to the center of the data scale in the saved image.  So, it centered the modulus values of the image at 0 MPa when the image was saved.  This is why your modulus data is offset so that you are getting negative values.  When you are collect the DMTModulus data channel (and the other PeakForce QNM channels such as Adhesion, Deformation, Dissipation, and LogDMTModulus), setting the Offline Planefit (OL Planefit) parameter to "None" will prevent this problem and save the raw data values.

Hope this helps.

John

  • | Post Points: 13

All Replies

Top 10 Contributor
60 Posts
Points 651

Thirst thing coming to mind: You are flattening the picture. check if the flatening is off when you record ( not on full after recording), also don't flatten after saving.

But perhaps i am wrong..... Zip it!

  • | Post Points: 10
Top 25 Contributor
29 Posts
Points 341
Bruker Employee
Verified by KAU_ECE

When you collected your data, the Offline Planefit parameter (called "OL Planefit" in the software) was set to "Offset."  This will offset the average Z value of the image to the center of the data scale in the saved image.  So, it centered the modulus values of the image at 0 MPa when the image was saved.  This is why your modulus data is offset so that you are getting negative values.  When you are collect the DMTModulus data channel (and the other PeakForce QNM channels such as Adhesion, Deformation, Dissipation, and LogDMTModulus), setting the Offline Planefit (OL Planefit) parameter to "None" will prevent this problem and save the raw data values.

Hope this helps.

John

  • | Post Points: 13
Top 200 Contributor
5 Posts
Points 56
KAU_ECE replied on Wed, Jan 15 2014 10:00 AM

Thank you for the reply. Yes, OL Planefit was set to Offset and when we turned it off, we are now able to see the raw data and all values are positive.

 

But sometimes we get streaks in the DMT Modulus image even when the height image is spotlessly clean. Which parameter should be adjusted to remove these streaks?

 

Height image:

 

How can we get a clean image as shown on page 3 (fig. c) of this brochure? https://www.bruker.com/fileadmin/user_upload/8-PDF-Docs/SurfaceAnalysis/AFM/Brochures/PeakForce_Quantitative_Nanomechanical_Property_Mapping_b.pdf 

  • | Post Points: 12
Top 10 Contributor
60 Posts
Points 651

Three things you should check:

  1. All your values are on the high part of your scale 200GPa and more and even more than the 320 GPa. Be sure that you use a probe that is stiff enough to measure this values.
  2. Be sure that your scale in the modulus settings is not set from 0 to 320 GPa but eg from 200 to 500 GPA so make sure your scaling is ok
  3. Make sure that you calibrate the tip ( Sens. deflection) on a very hard surface, but I think you did that

  • | Post Points: 12
Top 200 Contributor
5 Posts
Points 56
KAU_ECE replied on Wed, Jan 15 2014 11:48 AM

We are using the TAP150A probe. I had calibrated the deflection sensitivity on the glass side before starting the imaging.

 

In another slide, we got this image:

 

But even here we don't see a sharp contrast between the DMT Modulus readings of the fibers and the background.

  • | Post Points: 10