The Nanoscale World

error of the X,Y offset?

rated by 0 users
Answered (Not Verified) This post has 0 verified answers | 1 Reply | 1 Follower

Top 150 Contributor
6 Posts
Points 72
lmign posted on Mon, Jul 18 2011 2:41 PM

Hi All,

We are trying to measure the distance between two features in Dimension ICON, sometimes the distance between them could be large ( around 50micron) so it will be time consuming and not accurate enough to just scan a large area ( since the maximum lines/sample we can use is 5120).

So we are thinking to measure one feature first by a scanning a much smaller area and then set the x, y offset to move the tip to another feature and do the scanning

again (in closed loop). This method seems works but we are not sure how accurate the tip position was when we set the offset.

 

Besides, between closed loop and adaptive, which one will be better for us in terms of measuring distance between two far away features?

Thanks a lot.

 

 

 

  • | Post Points: 12

All Replies

Answered (Not Verified) replied on Mon, Jul 18 2011 3:04 PM

Hi Imign,

I am not sure about the accuracy you can ultimately achieve using the offset procedure that you envision. It should be relativelty straightforward for you to check it though. As for the closed-loop vs adaptive. The adaptive scan mode avoids hitting resonances and thus helps lower the sensor noise. ICON sensors are already so quiet that the effect is bascially negligable and therefore I recommend just continuing in closed-loop mode.

  • | Post Points: 10