Hi, All, we have a multimode 8 and we have purchased some DLCS probes from Bruker to image our nanowire samples. We expected to see a 2-3nm nanowire with this 1nm radius tip but the results showed much larger as 6-8nm. I was told by Bruker engineer we prob has got a blunt tip already upon engage and we need to tune engage parameters to preserve the sharpness of the supersharp tips. Can anyone provide a detailed procedure and tips how we should do when using a supersharp probe if it's so easy to get blunt even before we start to image? great thanks in advance!
Mikromasch produce similar probes, and have a set of engage parameters here (approx half way down the page):
http://www.spmtips.com/howto/res/hr
I use similar parameters on a multimode with sharp tips, however I use a lower engage gain (0.1) and also set the pre-engage setpoint to 95%. The most important parameter is "sew tip" this MUST be set to "yes". See the multimode or command reference manual for the nanoscope software for an explanation of the engage parameters and hints for optimizing them. Bear in mind that the engage will be a lot slower than normal with more cautious engage parameters.
Be aware that these tips are extremely fragile, so scanning can damage them too. For best results, you should tune to a relatively small free amplitude (this will depend on the sample and cantilever properties, but ~0.5V is a good starting point) and control at the highest setpoint that still allows you to track the surface. Keep the gains high, and scan slowly. Avoid large scans (more than a few hundred nm) and large amounts of topography - it can be helpful to scan a large area with a regular tip first to find a clean, flat area with features of interest, and then switch to a sharp probe, engaging straight onto the area of interest.
Also, as the carbon spikes are extremely hydrophobic, they are likely to pick up hydrophobic molecules from the surface, so your sample prep must be very clean.
Hope that helps.
Best wishes,
Nic
Many thanks, Nic, your tips very helpful.
Can any Bruker specialist explain the difference between soft tapping and standard tapping? and I tried small amlitude tapping as someone suggested it would help in terms of resolution, but i could not get it to work when I tuned the amplitude below 2nm, anyone have some tips on that? Thanks very much in advance!
I hope that helps!
--Bede